this is very confusing for me.
I was under the impression that with the latest US firmware every AirSense 10 AutoSet has the the "normal" AutoSet-Mode with the 2 responses and also the For-Her-Mode?
Meaning: the actual For-Her-device differs only in coloring.
Is this no longer the case - or maybe never was the case at all?
(the latest european firmware however does NOT enable the For-Her-Mode on the normal AutoSet - thus RERA-reporting only only the actual For-Her-Device)
Quote:The auto set for her Caps the EPAP at 12cm. So if you require a higher pressure you want to use the normal auto set mode.
that is not correct - as stated in the (clinical) manual for the For-Her-Mode.
In the For-Her mode the device only RESPONDS to
obstructive Apneas until 12cmH2O - it reacts however still to flowlimitations (and I'm unsure about RERAs - but I suppose it reacts to those as well above 12cmH2O)
And in For-Her-Mode it raises the minimum pressure for that session if 2 OAs occur within short range (I believe 1 or 2 minutes) - the minimum pressure should than be the resulting pressure after the 2nd OA (but you better look that up, if you want to know that exactly)
All in all your pressure will be much higher with the ResMed - compared to Philips - as ResMed mainly responds to flowlimitations. For-Her produces an overall lower pressure-need as the normal AutoSet-algorithm would produce.
I spare you the details about what or what little ResMed flags as events - just know that you will be seeing much less events (as compared to nearly every other brand) - this does not mean, that they are not there.